I have one DP transmitter with diaphragm capillary seal, the ordering number for Seal is 1199DA356AFCS74VI00ZT . i try to find the detail configuration of ordering number, but in new document revision is meaningless. can you help me? the transmitter is built in 2000 (i think)
Thanks
Hello, Your Model Number 1199DA356AFCS74VI00ZT can be broken down as follows: 1199 = 1199 Seal System D = Balanced A = Syltherm 3 = 2mm 56= 3 meter A = ANSI FCS = RTJ Flange 7 = 3” 4 = 600# VI = 2205 Duplex body, Alloy C-276 diaphragm (one piece) Z = Material certs to DN10204 DIN 3.1B T = NACE FR-01-75 As there have been some minor updates to our 1199 model options, the below model is an equivalent system and will need to be used on future orders: Updated Model Number: 1199DAD56AFCW74V100Q15E 1199 = 1199 Seal System D = Welded-Repairable system A = Syltherm XLT fill fluid D = 0.075 in. (1.905mm) capillary ID 56 = 3m capillary length A = ANSI FCW = Ring Type Joint Flange, Flush Surface 7 = 3in. process connection size 4 = class 600# flange V1 = Alloy C276 Diaphragm, Duplex 2205 Upper Housing, 2205 Duplex Mounting Flange 0 = No Flushing Connection 0 = No Lower Housing Q15 = Certificate of Compliance to NACE MR0175/ISO 15156 for wetted materials E = One Piece Design In order to get the Material Traceability Certification per EN 10204 3.1B (your “Z” option code), you will need to add the “Q8” option to the Rosemount transmitter model number that will be attached to the 1199 system. This certificate will cover both the transmitter and 1199 models Please let me know if you need any additional assistance. Best Regards, Ryan Kees, Sr. Product Manager, Emerson Process Management