Hi Musabthelion,
Typically having the analyzer not powering up could be for the following reasons..
1.- No 24VCD Present(please double check)
2.-Open Fuse
3.- 24 VCD Terminal not seated correctly on the receptacle.
4.- LOI Board and CPU not seated correctly.
5.- Backplane Board Damaged.
Note: Please make sure all connectors are correctly inserted.
By The Way , please provide instrument sales order and serial number of your GC370XA so I can check the factory records
Kind Regards.
Ray
Thank you,
Raymundo Morin | Global Training Manager | Rosemount Gas Analysis
Emerson Automation Solutions | 10241 West Little York, Suite #200 | Houston | TX | 77040 | USA
T +1 | M +1 346-773-1367 | F +1 713 396 8702 | Raymundo.Morin@emerson.com
Please guide about following,
Thanks and Regards,
MUSAB SHABBIR
In reply to musabthelion: