Oxygen Analyzer.

Can any Body tell me what is the Best Oxygen Analyzer currently Rose mount produce. 

6 Replies

  • Muhammed,

    This is too broad a question to answer . You have state the application and other process details
    Rosemount manufacture three dectectors /methods specific to oxygen measurement as well
    as GC ( gas chromatograph)and QCL ( Quatumn Cascade Laser )that can measure Oxygen.

    Sensor/ Detection Method:
    1. Electro Chemical ( Electrolyte Cell) -- Extractive , consumable electrolyte , life is dependent on level
    of O2 measured.
    2. Zirconia Oxide : Insitu , high temperature f0r flue gas . Life can be years or months depending
    on the process.
    3. Paramagnetic Method: Electro mechanical cell , responds to O2 based on magnetic
    susceptibility of O2 .

    Search on Analytical web site for Model 6888/ 4000xtractive , Process Gas analyser , X-Stream fro associated data sheets.
  • In reply to Jim McLelland:

    Mohammed , typing error 6888/ 4000 should read in-situ not extractive . Extractive methods apply to the PGA( Electro-chemical cell Paramagnetic methods) , GC and QCL.
  • 2. Zirconia Oxide : Insitu , high temperature f0r flue gas . Life can be years or months depending
    on the process.
    Dear Jim;


    Thank you for your reply, kindly consider the Zirconia OXide base sensor for the selection of Analyzer.
  • In reply to Muhammad Yesrub Iqbal:

    Muhammad ,

    My information was for your reference only. If you need to follow up with a view to purchasing a suitable analyser please contact your local Emerson/ Rosemount office.
    I am not located in a sales office . You have not entered any profile , best you contact a sales office to further process your requirement.
  • In reply to Muhammad Yesrub Iqbal:

    Hi Muhammad,

    We have several product offerings that utilize zirconia oxygen sensors. Below is some information about our products. I highly recommend you contact your local sales office or email [email protected] to get in touch with a representative to further assist you based on your application requirements.

    Oxymitter: In situ oxygen probe and transmitter available for general purpose or hazardous location applications. Suitable for high temperature applications when used in conjunction with a bypass,.
    6888: In situ oxygen probe and transmitter available for general purpose applications only. Includes addition options including integrated autocalibration and advanced process features. Suitable for high temperature applications when used in conjunction with a bypass
    OCX8800: Close-coupled extractive oxygen and combustibles transmitter. Suitable for standard and high temperatures
    5081FG: In situ oxygen probe and transmitter for high temperature applications only. Non-repairable.

  • In reply to Muhammad Yesrub Iqbal:

    Let me know if you want any info on these models, we have used most models of ZrO2 Emerson models in a variety of services.
    Not detracting from the Emerson reps here, of course.