Where can i get  the complete Modbus Register Table for the 3095FC Multivariable Flow Transmitter?

  • Steve,

    The Rosemount 3095FC is a discontinued product, but the Product Manual is still available on Emerson.com.  Here is a direct link to the Product Manual: https://www.emerson.com/documents/automation/manual-rosemount-3095fc-en-76020.pdf

    The Product Manual available at that link provides information on Modbus registers on page 3-47 and in Appendix C.

    If you are not able to find the information needed in the product manual, please send me an email at Nathan.Stokes@Emerson.com and I will connect you with our technical support team to get your question answered.

    Thank you,

    Nathan

  • Thanks Nathan,
    Info that I needed.
     
    Steve Johnson