  • Good day,

    We have the problems with pressure transmitters 2051C-G-1-A-2-2-A-1-B-E1-M5-D4-DF with measuring range -50…+50 mbar mounted on tanks (insulated and not). When the weather become colder some of them start to indicate vacuum, sometimes the measurements drops to LRV. Sensor zero trim doesn't help. After some time the pressure readings return to normal. What could be the problem?
  • In reply to alekseib:

    Hello - What are you measuring and how low is the temperature getting down to? and is it possible that it is freezing in the transmitter? Thanks Wally
  • In reply to alekseib:

    In addition to the suggestion Wally made, another possibility is that you have a blanket pressure in the tank that is not being controlled properly, which in the application you describe should be a constant pressure. You mentioned that the tanks are insulated, which would suggest that these are closed tanks. When you use a GP transmitter to measure the level in the tank, any fluctuation in the pressure inside a tank would be seen as a change in level inside the tank. We would typically measure a closed tank with a DP transmitter, versus a GP transmitter (as indicated by the G in the model code). Using a DP transmitter allows accurate measurement of the level in the tank regardless of any fluctuations in the blanket pressure.

     

  • In reply to alekseib:

    There's also a possibility that the low side gage vent could be plugging temporarily due to rain or freezing rain. When it warms up either the ice melts or the rain dries up and the gage vent then reopens. This transmitter is a gage style with a coplanar flange which has no process connection on the low side with 3 vent paths. A suggestion would be to shield the transmitter with something to see if the issue goes away if you think that it’s a water issue. You could use something as simple as a plastic bag.
  • In reply to alekseib:

    Hello Friendly reminder to join tomorrow at 1:00pm Singapore time when the experts pictured above will be here LIVE to answer all of your toughest Measure & Analyze questions...

    Best Regards,

    Rachelle McWright: Business Development Manager, Dynamic Simulation: U.S. Gulf Coast

  • In reply to Rachelle McWright:

    i like to know if Nitrogen can be used to zero Zirconia Oxygen Analyzer?
  • In reply to Audrey Seck:

    The base answer is no . N2 will give you an excellent zero point however the N2 will also create a very long cell recovery time as you introduce the process or calibration gas . We can discuss the reason for this offline.
  • Comparing a temperature transmitter versus a wired direct temperature sensor, what are the differences in performance?
  • In reply to Jim Cahill:

    Converting a low-level sensor signal to a high-level signal as soon as possible, as close to the sensor as possible is best practice. If you are looking for minimal cost, consider using multiple input temperature transmitters such as 4-input wireless or 8-input fieldbus temperature transmitters
  • What type of diagnostic information is available from radar level transmitters?
  • This one came in from LinkedIn, "If all calculations that done in Flow Computer can be done in DCS then why we need a dedicated Flow Computer for Fiscal metering?"
  • In reply to Jim Cahill:

    The reason you need a flow computer instead of DCS is basically for security and fidelity. The flow computer carries out the flow calculations according to international standards whereas the DCS only has limited calculations capacity and is less secure than a flow computer.
  • Why is Hydrogen used as Carrier Gas in Gas Chromatography?
  • Is radar level technology suitable in custody transfer applications?