hi i have problem with micro motion f300 transmitter 2700 showing code A21 , the K1 value from sensor tag also I checked the sensor type (curved).
please help to solve this problem.
Hi Ahmed,
The most common reason for this error is that the characterization is incorrect. Make sure you haven't entered values into any of the straight tube fields. You should check that the following items say the same as what is printed on the sensor tag:
-Flow calibration factor
-D1 and D2
-DTC or DT
-K1 and K2
-FD
Here is a video that steps through the configuration and characterization process, too, that might be helpful: http://videos.emersonprocess.com/detail/videos/micro-motion/video/5519215366001/micro-motion-coriolis-meter-characterization-and-configuration-using-prolink?autoStart=true&page=2
The characterization instructions start at 2:15 in the video.
Best Regards, Tonya
In reply to Tonya Wyatt:
In reply to ahmed bekhta:
It sounds like you have the device configured correctly. I would recommend sending an e-mail to [email protected] and they can help you with full troubleshooting to get the device working correctly.
Best, Tonya