Typically having the analyzer not powering up could be for the following reasons..
1.- No 24VCD Present(please double check)
2.-Open Fuse
3.- 24 VCD Terminal not seated correctly on the receptacle.
4.- LOI Board and CPU not seated correctly.
5.- Backplane Board Damaged.
Note: Please make sure all connectors are correctly inserted.
By The Way , please provide instrument sales order and serial number of your GC370XA so I can check the factory records